FDA MedWatch - Maquet Datascope Intra-Aortic Balloon Pumps: Class I Recall - Failure of Fan Assembly

[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]

 



Title: FDA MedWatch - Maquet Datascope Intra-Aortic Balloon Pumps: Class I Recall - Failure of Fan Assembly
MedWatch logoMedWatch - The FDA Safety Information and Adverse Event Reporting Program

Maquet Datascope Intra-Aortic Balloon Pumps: Class I Recall - Failure of Fan Assembly

Including System 98/98XT, CS100/CS100i, and CS300, Model Numbers:  0998-00-0446-xx, 0998-UC-0446-xx, 0998-00-0479-xx, 0998-UC-0479-xx, 0998-UC-0446Hxx, 0998—00-3013-xx, 0998-UC-3013-xx, 0998-00-3023-xx, 0998-UC-3023-xx; and various lot numbers

AUDIENCE: Risk Manager, Cardiology, Biomedical Engineering

ISSUE: FDA notified health professionals of a Class I Recall due to a fan assembly for the affected Intra-Aortic Balloon Pumps (IABP) that could contain a misshapen retaining ring. This retaining ring could separate within the fan assembly, causing the fan to stop rotating. This could result in the power supply overheating and the IABP would shut down without any warning.

This device failure would cause the balloon to stop inflating and deflating. It may result in decreased blood flow to the heart and the rest of the body, difficulty in weaning the patient from cardiopulmonary bypass, and clotting or blockage of blood vessels to the intestines, kidneys, or legs. The possible long range consequences of this include organ injury or tissue damage, possibly leading to patient death.

Affected products were manufactured From January 1, 2003 through June 30, 2011 and distributed from January 17, 2003 through June 30, 2011.

BACKGROUND: An intra-aortic balloon pump is a mechanical device that inflates a balloon placed in the patient’s thoracic aorta. The balloon inflates and deflates in order to help the heart pump blood, increase blood flow to the heart itself, and increase blood flow to the rest of the body.  The device is used temporarily for emergency cardiac support conditions, such as after a heart attack, for severe heart failure, or while the patient is waiting for heart surgery.

RECOMMENDATION: On March 21, 2014, Maquet sent an Urgent Medical Device Field Correction letter to all affected customers. The letter identified the product, the problem, and the actions to be taken.

A recalled pump may continue to be used until a repair or replacement is performed. However, there should be heightened awareness that a sudden pump failure may occur without warning, and a backup pump should be available for immediate substitution in the case of a failure. If an alternative pump is unavailable and a failure occurs, the balloon should be removed from the patient as soon as possible according to standard procedures.  This could result in added risk to the patient.

A Maquet Service Representative will contact customers with the affected IABPs to replace the fan assembly. For customers with technical questions within the U.S., contact Technical Support at 1-800-777-4222 and press 3, Monday through Friday, between 8:00 a.m. and 6:00 p.m., Eastern Daylight Time.

Read the MedWatch safety alert, including a link to the FDA Recall Notice, at:

http://www.fda.gov/Safety/MedWatch/SafetyInformation/SafetyAlertsforHumanMedicalProducts/ucm396903.htm

 


This email was sent to list-fda@xxxxxxxxxxx using GovDelivery, on behalf of: U.S. Food & Drug Administration (FDA) · 10903 New Hampshire Ave · Silver Spring, MD 20993 · 800-439-1420 Powered by GovDelivery

[Index of Archives]     [CDC News]     [NIH News]     [USDA News]     [Steve's Art]     [Camping in Yosemite]     [PhotoForum]     [SB Lupus]     [STB]

  Powered by Linux