GE Healthcare Resuscitation Systems: Class I Recall - Oxygen and Air Inlet Fittings Reversed During Assembly

Includes the following models:

 

AUDIENCE: Risk Manager, Biomedical Engineer

ISSUE: The oxygen and air wall inlet fittings on the back panel of the resuscitation systems were reversed during assembly. These recalled products may interfere with oxygen delivery,  resulting in inaccurate oxygen regulation in newborns (neonates), and may lead to low blood oxygen (hypoxia) or high blood oxygen (hyperoxia). This may cause death in neonates, particularly those who are critically ill. Preterm and low-birth weight babies are also at increased risk in terms of morbidity and mortality.

The affected products were manufactured from April 1, 2007 through October 31, 2013 and distributed from October 1, 2007 through October 31, 2013. See the Recall Notice for listing of affected lot numbers.

BACKGROUND: On November 18, 2013 GE Healthcare sent an Urgent Medical Device Correction letter, with a return receipt, to all affected customers. The letter identified the products, the problem, and safety instructions to be taken by the customer along with the product correction. The firm recommends their customers alert all potential users in their facilities about the safety notification and the recommended actions.

RECOMMENDATION: The firm recommends their customers alert all potential users in their facilities about the safety notification and the recommended actions.

For questions or concerns regarding this recall, please call the company at 1-800-345-2700, Monday – Friday, 8:00 am – 4:30 pm.

Healthcare professionals and patients are encouraged to report adverse events or side effects related to the use of these products to the FDA's MedWatch Safety Information and Adverse Event Reporting Program:

Read the MedWatch safety alert, including a link to the recall notice, at:

http://www.fda.gov/Safety/MedWatch/SafetyInformation/SafetyAlertsforHumanMedicalProducts/ucm387731.htm


This email was sent to list-fda@xxxxxxxxxxx using GovDelivery, on behalf of: U.S. Food & Drug Administration (FDA) · 10903 New Hampshire Ave · Silver Spring, MD 20993 · 800-439-1420 Powered by GovDelivery

[Index of Archives]     [CDC News]     [NIH News]     [USDA News]     [Steve's Art]     [Camping in Yosemite]     [PhotoForum]     [SB Lupus]     [STB]

  Powered by Linux