FDA MedWatch - CT Brain Perfusion Scans Safety Investigation: Initial Notification

[Date Prev][Date Next][Thread Prev][Thread Next][Date Index][Thread Index]

 



Title: FDA MedWatch - CT Brain Perfusion Scans Safety Investigation: Initial Notification
MedWatch logo MedWatch - The FDA Safety Information and Adverse Event Reporting Program

Audience: Radiological, Neurological and emergency medicine healthcare professionals

[Posted 10/09/2009] FDA notified healthcare professionals that it has become aware of radiation overexposures during perfusion CT imaging performed to aid in the diagnosis of stroke at a particular facility, the patients receiving radiation doses that were approximately eight times the expected level. While this event involved a single kind of diagnostic test at one facility, the magnitude of these overdoses and their impact on the affected patients were significant. This situation may reflect more widespread problems with CT quality assurance programs and may not be isolated to this particular facility or this imaging procedure (CT brain perfusion). If patient doses are higher than the expected level, but not high enough to produce obvious signs of radiation injury, the problem may go undetected and unreported, putting patients at increased risk for long-term radiation effects.

FDA encourages every facility performing CT imaging to review its CT protocols and be aware of the dose indices normally displayed on the control panel. These indices include the volume computed tomography dose index and the dose-length product. For each protocol selected, and before scanning the patient, carefully monitor the dose indices displayed on the control panel. To prevent accidental overexposure, make sure that the values displayed reasonably correspond to the doses normally associated with the protocol. Confirm this again after the patient has been scanned. Patients should follow their doctor’s recommendations for receiving CT scans. While unnecessary radiation exposure should be avoided, a medically-needed CT scan has benefits that outweigh the radiation risks.

Read the MedWatch safety summary, including a link to the FDA Initial Notification, at:

http://www.fda.gov/Safety/MedWatch/SafetyInformation/SafetyAlertsforHumanMedicalProducts/ucm186105.htm 


You are encouraged to report all serious adverse events and product quality problems to FDA MedWatch at www.fda.gov/medwatch/report.htm

Update your subscriptions, modify your e-mail address, or stop subscriptions at any time on your Subscriber Preferences Page. You will need to use your e-mail address to log in. If you have questions or problems with the subscription service, please contact support@xxxxxxxxxxxxxxx.

This service is provided to you at no charge by U.S. Food & Drug Administration (FDA).

FDA HHS

GovDelivery, Inc. sending on behalf of U.S. Food & Drug Administration (FDA) · 5600 Fishers Lane · Rockville MD 20857 · 800-439-1420